6.3.3 Completion of the measurements
Before unplugging the Measurement cable from the device receptacle, disconnect the Lip
Clip and the File Clip from the patient.
Move the file stopper to the selected reference point on the tooth.
Gently remove the file from the canal and measure the apical length between the stopper
and the file tip.
Determination of working length for canal shaping is a subject of dentist's professional
judgment. In most cases subtraction of 0.5 mm from the measured apical length provides
clinically acceptable working length. Nevertheless, in each case the dentist should define
proper working length basing on his experience, apex locator readings, radiographs and
other available data.
6.4
Audio feedback
Wirele-X is equipped with an audio indicator which accompanies file progression within
the canal in the pre-apical and apical zones. This function is activated in parallel with the
LEDs and enables monitoring of file advance in the apical zone, even without seeing the
device.
6.5
Automatic Shutdown
Wirele-X automatically shuts down after 5 minutes without use. In order to prolong the
battery life, it is recommended to switch off the device after completing the measurements
by pressing the On/Off button.