Wirele-X is intended for use in the electromagnetic environment specified below;
The customer or the user of Wirele-X should assure that it is used in such an environment.
Immunity test
Rated power
frequency
magnetic fields
IEC 61000-4-8
Note:
• UT is the a.c. mains voltage prior to application of the test level.
Conducted
3 Vrms
disturbances
inducted by RF
150 kHz to 80 MHz
fields
6 Vrms in ISM
bands 150 kHz to
80 MHz
80% AM at 1 kHz
IEC 61000-4-6
10 V/m
Radiated RF
80 MHz to 2,7 GHz
IEC 61000-4-3
Guidance and manufacturer's declaration:
electromagnetic immunity - Wirele-X
Immunity test
30 A/m
50 or 60 Hz
Immunity test
30 A/m
50 or 60 Hz
3 Vrms
150 kHz to 80 MHz
6 Vrms in ISM bands
150 kHz to 80 MHz
80% AM at 1 kHz
10 V/m
- 18 -
Immunity test
Power frequency magnetic fields
should be at levels characteristic
of a typical public low-voltage
power supply network that
supplies buildings used for
domestic purposes, commercial or
hospital, clinic environment.
Portable and mobile RF
communications equipment should
be used no closer to any part of
Wirele-X, including cables, than
the recommended separation
distance calculated from the
equation applicable to the
frequency of the transmitter.
Recommended separation distance
d = 1.17√P
d = 1.17√P
80 MHz to 800 MHz
d = 2.3√P
800 MHz to 2,7 GHz
Where P is the maximum output
power rating of the transmitter in
watts (W) according to the
transmitter manufacturer and d i s
the recommended separation
distance in meters (m).